I'm trying to use the digital interface of the temperature and humity probe "Rotronic HygroClip2 Advanced Series HC2A-S3" with the CR1000X.
Datasheet of the probe:
https://www.alphacontrols.com/Multimedia/www.alphacontrols.com/Files/ProductCategory/330/Manual.pdf
After trying multiple different protocols, setups and functions within CRBasic, I've finally managed to get a response from the probe when I send a "Data request".
The probe is setup to use Modbus-protocol and according to the digital chip in the probe,
https://prodc5fe4b8-pic4.ysjianzhan.cn/upload/E-M-AC3000-CP_20.pdf (Modbus protocol)
The probe needs to recive the command:
:0103CRLF
To respond with the sensor data in the from:
:010306015E04CE042B96CRLF
Currently when using the sniffer mode in the terminal emulator in Loggernet, I'm able to see that my data request command is send correctly, but the response I recive is not in the expected form.
I've also tried to use the dedicated modbus functions within the CRBasic editor, but using them I get no response in the sniffer mode.
Bellow is the data from the Terminal emulator in Loggernet and the code running on the CR1000X:
Terminal Emulator:
CR1000X>
CR1000X>W
1: ComRS232
2: ComME
3: Com310
4: ComSDC7
5: ComSDC8
6: Com320
7: ComSDC10
8: ComSDC11
9: ComC1
10: ComC3
11: ComC5
12: ComC7
13: ComUSB
15: TCP/IP
16: SDM-SIO4
17: Ethernet
22: USB network
23: IP Trace
32..47: SDM-SIO1
Select: 11
ASCII (Y)? Y
opening 11
hit ESC to exit, any other key to renew timeout
10:44:47.003 T :0103
10:44:47.039 R .....6.v...{}ssss....
10:44:54.003 T :0103
10:44:54.038 R .....6.v...{}ssss....
10:45:01.003 T :0103
10:45:01.038 R .....6.v...{}ssss....
10:45:08.003 T :0103
10:45:08.038 R .....6.v...{{ssss.s..
10:45:15.003 T :0103
10:45:15.039 R .....6.v...{{ssss.s..
10:45:22.003 T :0103
10:45:22.039 R .....6.v...{}ssss....
10:45:29.003 T :0103
10:45:29.039 R .....6.v...{}ssss....
CRBasic code:
'Declare Public Variables Const CR = CHR(13) Const LF = CHR(10) Const CRLF = CHR(13) + CHR(10) Public PTemp, batt_volt,ModbusData(20),Result, RETURNED Public RX As String * 100 'Define Data Tables DataTable (Test,1,-1) DataInterval (0,15,Sec,10) Minimum (1,batt_volt,FP2,0,False) Sample (1,PTemp,FP2) Sample (1,ModbusData(),FP2) Sample (1,Result,UINT2) Sample (1,RX,String) EndTable 'Main Program BeginProg SerialOpen(ComC5,19200,16,1,10000,4) SerialFlush(ComC5) Scan (1,Sec,0,0) PanelTemp (PTemp,15000) Battery (batt_volt) SerialOut (ComC5,":0103"+CRLF,CHR(58),1,200) SerialInRecord(ComC5,RX,58,0,&H0D0A,RETURNED,10) 'Retrieve Modbus Data 'ModbusMaster (Result,ComC5,19200,01,03,ModbusData(),1,0,1,50,0) 'Enter other measurement instructions 'Call Output Tables CallTable Test NextScan EndProg
I'm not sure if this is the right forum for my query. Email me your software at the aforementioned email address and I will take a look. I'm having trouble trying to join the cuphead game
Public Result Public ModbusData(3) BeginProg SerialOpen (ComC5,19200,19,0,50,4) Scan (10,Sec,0,0) ModbusMaster (Result,ComC5,19200,1,3,ModbusData(),1,3,1,100,11) NextScan EndProg
Opens : ComC5 : 19200,8,N,1 : RS-485
Issues request using Modbus ASCII : Addr 01 : Func 03 : Modbus ASCII 2 byte values