Our full technical support staff does not monitor this forum. If you need assistance from a member of our staff, please submit your question from the Ask a Question page.


Log in or register to post/reply in the forum.

Polling Vaisala PTB330 over RS232 with a CR6 Datalogger.


RJP Jul 12, 2025 04:04 PM

Am attempting to interface a Vaisala PTB330 over RS232 to a CR6 datalogger via the user terminals.

For clarity, the PTB330 is wired to the CR6 as such

PTB330 - User Port RS232 TX to CR6 C2

PTB330 - User Port RS232 RX to CR6 C1

PTB330 - User Port RS232 GND to CR6 GND

_________________________


The PTB330 Serial settings are:

9600b/s

8N1

Poll Adress:0

________________________

I've confirmed the unit is outputting barometric pressure as an ASCII string in a 6 digit format with 2 decimal places via connecting directly with Putty ie. 

1014.15

1014.18


I am having issues with converting this data to a floating point number to be logged within the unit as currently the output to the CR6 registers as NAN(Not a number). 

At first I wanted to try and read it as a string, but I'd prefer the data stored directly as a floating point into a table for ease of use.


I've tried a few different approaches with trying to store the string inside a variable then convert it to a float before outputting it in the table. However I cannot seem to get it to work.

 

'CR6 Series
'Created by Short Cut (4.7)

'Declare Variables and Units
Public BattV
Public PTemp_C
Public AirTC
Public RH
Public WS_ms
Public WindDir
Public RawTempV
Public RawHumidV

Public RawBuffer      As String * 50
Public CleanPressure  As String * 20 

Units BattV       = Volts
Units PTemp_C     = Deg C
Units AirTC       = Deg C
Units RH          = %
Units WS_ms       = meters/second
Units WindDir     = degrees

'Define Data Tables
DataTable(Hourly, True, -1)
    DataInterval(5, 60, Sec, 10)
    MQTTPublishTable(1,0,10,Min,2,0,0,0)
    Minimum(1, BattV,      FP2, False, True)
    Maximum(1, BattV,      FP2, False, True)
    Minimum(1, PTemp_C,    FP2, False, True)
    Maximum(1, PTemp_C,    FP2, False, True)
    Minimum(1, AirTC,      FP2, False, True)
    Maximum(1, AirTC,      FP2, False, True)
    Minimum(1, RH,         FP2, False, True)
    Maximum(1, RH,         FP2, False, True)
    Maximum(1, WS_ms,      FP2, False, True)
    Minimum(1, WS_ms,      FP2, False, True)
    Sample (1, WindDir,       FP2)
    Sample (1, CleanPressure, String)
EndTable

DataTable(Seconds, True, -1)
    DataInterval(5, 60, Sec, 10)
    MQTTPublishTable(1,0,10,Min,2,0,0,0)
    Minimum(1, BattV,      FP2, False, False)
    Sample (1, AirTC,      FP2)
    Sample (1, RH,         FP2)
    Sample (1, CleanPressure, String)
EndTable

'Main Program
BeginProg

    SerialOpen(COMC1, 9600, 0, 0, 25)

    'Main Scan
    Scan(5, Sec, 1, 0)
        Battery(BattV)
        PanelTemp(PTemp_C, 60)

'Poll PTB330 & Read ASCII SerialOutBlock(COMC1, "SEND", 4) Delay(0, 500, mSec) SerialInRecord(COMC1, RawBuffer, &H20, 0, &H0D, RawBuffer, 1) If RawBuffer <> "" Then ' The idea here is to trim and convert to float, then format to "XXXX.XX" Sprintf(CleanPressure, "%06.2f", Float( Trim(RawBuffer) )) EndIf 'MP100A Temp & RH Sensor VoltSE(AirTC, 1, mV1000, U1, False, 0, 60, 0.1, 0) VoltSE(RawTempV, 1, mV1000, U1, False, 0, 60, 1, 0) VoltSE(RH, 1, mV1000, U2, False, 0, 60, 0.1, 0) VoltSE(RawHumidV, 1, mV1000, U2, False, 0, 60, 1, 0) If (RH > 100) AND (RH < 108) Then RH = 100 '05103 Wind Sensor PulseCount(WS_ms, 1, U6, 5, 1, 0.098, 0) BrHalf(WindDir, 1, mV5000, U4, U3, 1, 2500, True, 20000, 60, 355, 0) If (WindDir >= 355) OR (WindDir < 0) Then WindDir = 0 CallTable Hourly CallTable Seconds NextScan EndProg

The 

Sprintf(CleanPressure, "%06.2f", Float( Trim(RawBuffer) )) 

doesn't feel like the right way to do it. 


Any assistance would be greatly appreciated.


JDavis Jul 14, 2025 08:22 PM

Do a SerialFlush() before the SerialOutBlock() to empty the serial buffer of anything left over.

If the output is fixed width, you can change SerialInRecord() by setting BeginWord parameter to 0 and number of expected bytes to 7. Otherwise, use SerialIn() instead of SerialInRecord(). Destination variable should be a string.

SplitStr with option 0 should pull the floating point number out of the string.


RJP Jul 17, 2025 03:56 PM

Thanks JDavis for your assistance.

Updated CRBasic based on your advice. 

'CR6 Series

'Declare Variables and Units
Public BattV
Public PTemp_C
Public AirTC
Public RH
Public WS_ms
Public WindDir
Public RawTempV
Public RawHumidV
Public NBytesReturned
Public FP_Pressure As String *20

Units BattV           = Volts
Units PTemp_C         = Deg C
Units AirTC           = Deg C
Units RH              = %
Units WS_ms           = meters/second
Units WindDir         = degrees
Units FP_Pressure     = hPa

'Define Data Tables
DataTable(Hourly, True, -1)
    DataInterval(5, 60, Sec, 10)
    MQTTPublishTable(1,0,10,Min,2,0,0,0)
    Minimum(1, BattV,        FP2, False, True)
    Maximum(1, BattV,        FP2, False, True)
    Minimum(1, PTemp_C,      FP2, False, True)
    Maximum(1, PTemp_C,      FP2, False, True)
    Minimum(1, AirTC,        FP2, False, True)
    Maximum(1, AirTC,        FP2, False, True)
    Minimum(1, RH,           FP2, False, True)
    Maximum(1, RH,           FP2, False, True)
    Maximum(1, WS_ms,        FP2, False, True)
    Minimum(1, WS_ms,        FP2, False, True)
    Sample (1, WindDir,      FP2)
    Sample (1, FP_Pressure,  FP2, False,)
    
EndTable

DataTable(Seconds, True, -1)
    DataInterval(5, 60, Sec, 10)
    MQTTPublishTable(1,0,10,Min,2,0,0,0)
    Minimum(1, BattV,          FP2, False, False)
    Sample (1, AirTC,          FP2
    Sample (1, RH,             FP2)
    Sample (1, FP_Pressure,    FP2)
EndTable

'Main Program
BeginProg
  
    'RS232 Setup
    SerialOpen(COMC1, 9600, 0, 0, 30)

    'Main Scan
    Scan(5, Sec, 1, 0)
        Battery(BattV)
        PanelTemp(PTemp_C, 60)
        
    'Poll PTB330 & Read ASCII
        SerialFlush(ComC1)
        SerialInRecord (ComC1,FP_Pressure,0,8,&H0A,NBytesReturned,01)
        Trim(FP_Pressure)   
        
        'MP100A Temp & RH Sensor
        VoltSe(AirTC,      1, mV1000, U1, False, 0, 60, 0.1, 0)
        VoltSe(RawTempV,   1, mV1000, U1, False, 0, 60,   1, 0)
        VoltSe(RH,         1, mV1000, U2, False, 0, 60, 0.1, 0)
        VoltSe(RawHumidV,  1, mV1000, U2, False, 0, 60,   1, 0)
        If (RH > 100) AND (RH < 108) Then RH = 100

        '05103 Wind Sensor
        PulseCount(WS_ms, 1, U6, 5, 1, 0.098, 0)
        BrHalf(WindDir,   1, mV5000, U4, U3, 1, 2500, True, 20000, 60, 355, 0)
        If (WindDir >= 355) OR (WindDir < 0) Then WindDir = 0

        CallTable Hourly
        CallTable Seconds
    NextScan
EndProg

 

I used the terminal emulator and have read the incoming from the port with the PTB330 setup to SEND a reading at 5Hz, the incoming from the unit with ASCII turned off in the terminal emulator is as follows.

21:07:46.607 R 31 30 32 30 2E 34 33 0D 0A                       1020.43..


21:07:47.607 R 31 30 32 30 2E 34 33 0D 0A                       1020.43..


21:07:48.606 R 31 30 32 30 2E 34 33 0D 0A                       1020.43..

With ASCII enabled - 

23:22:29.080 R 1016.23
23:22:34.086 R 1016.23
23:22:39.084 R 1016.23

I can see the string numbers in the Monitor Data tab now (hooray!) within PC400 however, when viewing the data log file in PC400, the data shows up as "1016.23". How do I have this entered into my tables as a FP number with 2 decimals instead of what seems to be a number string. 

I tried SplitStr however was receiving and error related to ResultVar(). I am not sure if I am using it correctly.

Thanks.

Log in or register to post/reply in the forum.